Gage R&R Study (Crossed)ANOVA Method |
Two-way ANOVA (analysis of variance) tells you if different levels of two factors result in different average values of a response variable.
The two-way ANOVA table lists the following sources of variability:
You want to see if the p-value associated with each source is less than the chosen level of significance. This would indicate that the source contributes significantly to the measurement variability.
Part 9 2.05871 0.228745 39.7178 0.000 Operator 2 0.04800 0.024000 4.1672 0.033 Part * Operator 18 0.10367 0.005759 4.4588 0.000 Repeatability 30 0.03875 0.001292 Total 59 2.24913
α to remove interaction term = 0.05 |
For the parts data, the p-value for Operator (P = 0.033) shows you that the average value of thickness is significantly different among the three different operators at the a = 0.05 level of significance.
Here, the Operator*Part term is included because P = 0.000 is less than 0.05. The interaction between operator and part will be taken into account in the Gage R&R.